HP-3468A, Display Voltage VDC
05-03-2021, 08:41 PM
Post: #23
RE: HP-3468A, Display Voltage VDC
(05-02-2021 07:24 PM)Dave Frederickson Wrote:  
(05-02-2021 07:16 PM)Garth Wilson Wrote:  The noise I mentioned being caused by water-soluble flux would also cause an unsteady reading. The head post indicated a steady reading of 2.86V though.

The 3468 is a 5.5-digit meter. The first post indicates a reading of 2.86XXX volts where XXX is the unsteady portion of the reading. Every 3468 exhibits this behavior.

My 3468, for example, displays 0.2XXXXX (0.3V range) and slowly creeps up to 2.XXXXX volts (3V range) with nothing connected over several minutes. So the whole display is "unsteady" when nothing is connected and works perfectly when provided an input.

Thank you Dave. The creeping is what my 3468 does. Up to 2,86XXX VDC. Not mV, not AC.
When nothing is connected.
Same as yours.
Find all posts by this user
Quote this message in a reply
Post Reply 


Messages In This Thread
HP-3468A, Display Voltage VDC - Uli - 04-23-2021, 04:26 PM
RE: HP-3468A, Display Voltage VDC - Uli - 04-23-2021, 06:03 PM
RE: HP-3468A, Display Voltage VDC - Uli - 04-24-2021, 12:32 PM
RE: HP-3468A, Display Voltage VDC - Uli - 04-24-2021, 04:34 PM
RE: HP-3468A, Display Voltage VDC - Uli - 04-25-2021, 05:11 AM
RE: HP-3468A, Display Voltage VDC - BobVA - 04-24-2021, 02:00 AM
RE: HP-3468A, Display Voltage VDC - Uli - 04-24-2021, 12:31 PM
RE: HP-3468A, Display Voltage VDC - Uli - 04-26-2021, 05:56 AM
RE: HP-3468A, Display Voltage VDC - cruff - 05-02-2021, 07:48 PM
RE: HP-3468A, Display Voltage VDC - Uli - 05-03-2021 08:41 PM
RE: HP-3468A, Display Voltage VDC - Uli - 06-01-2021, 07:39 PM



User(s) browsing this thread: 1 Guest(s)